Great! I will be there on Wednesday giving a course on Multiview reconstruction...hope to see you around!

All the best,
Stephan

Sent from my iPhone

On Aug 29, 2016, at 16:29, Tania Mendonca <tvmendonca1 at sheffield.ac.uk<mailto:tvmendonca1 at sheffield.ac.uk>> wrote:

Hi Neil and Stephan

Thanks for the tips and all the help! I tried again with a higher concentration of the 1µm beads, got rid of the bright big beads and increased the z acquisition distance to increase the region of overlap and was finally successful!

I hope to meet many of you OpenSPIMmers in Sheffield this week!

Best wishes
Tania

On 25 August 2016 at 15:08, Anthony, Neil <nantho2 at emory.edu<mailto:nantho2 at emory.edu>> wrote:
Hi Tania,

I'd agree with Stephan on the larger z stacks. I usually take >300 um, but it kinda depends on the field of view of the camera, as at some point you're taking images in an area that will no longer overlap.

400 interest points would be sufficient if they all overlapped, but more are required in general. I believe Stephan's articles suggest 1000-2000 to get the best results. All depends on overlap percentage.

When defining the beads I aim to use a sigma and threshold that picks out the peak of the bead if possible. Stephan, can I ask if it's detrimental when the threshold is too low and I'm identifying the same bead on many z planes? Also, I assume the sigma value determines the steepness of the point, but it still seems to pick up on larger aggregated beads etc, so I'm not too sure about that part.

Neil




________________________________
From: Stephan.Preibisch at mdc-berlin.de<mailto:Stephan.Preibisch at mdc-berlin.de>
Sent: Aug 25, 2016 3:41 AM
To: Tania Mendonca
Cc: Anthony, Neil; openspim at openspim.org<mailto:openspim at openspim.org>

Subject: Re: [OpenSPIM] Multiview registration - no inliers

Hi Tania,

from looking at the data I agree with Anthony, there are some problems which make it hard. But trying different calibrations and not getting a result makes me believe that there is maybe no overlap in between the stacks. You have quite a few of these big blobs apparently, so is it possible that these are different blobs?

If you run it again, make the stacks way bigger in z to make sure that there is overlap, increase the laser power, and maybe put some simple auto fluorescent sample in there so it is possible to judge if the calibration is right and if there is overlap.

Good luck,
Stephan
---

Dr. Stephan Preibisch
Group Leader

Berlin Institute of Medical Systems Biology of the MDC
Building 89, 1.08b

email: stephan.preibisch at mdc-berlin.de<mailto:stephan.preibisch at mdc-berlin.de>
web: http://www.preibisch.net/<http://fly.mpi-cbg.de/preibisch>

On 24 Aug 2016, at 19:49, Tania Mendonca <tvmendonca1 at sheffield.ac.uk<mailto:tvmendonca1 at sheffield.ac.uk>> wrote:

Hi Neil

Thanks for your email! The stacks should have the voxel size correctly set to 0.235 x 0.235 x 0.750. I'm assuming you've looked at the data on µManager as I'm unfamiliar with the ViewSetup tag that you mention. My system is not quite an OpenSPIM - I use LabView to control and don't have the Picard 4D stage (I use a Piezo for scanning instead).

Your manual transform looks about right, the stacks are about 120º apart with the big bead somewhere near the middle. I have tried a few different µbeads and the ones in these stacks are 1µm green beads but as you say they are nowhere near as bright as the big beads. Perhaps I do need to dump the large beads and find something else to try and reconstruct. I've been finding about 400-ish interest points in my stacks- is there a magic number that works for you? I've been imaging to about 100µm depth, how much would you recommend? Don't the stacks start getting too large to handle?

Thanks for the tip on the camera noise! I think I'll try again tomorrow with brighter beads, no big beads and larger volumes - I'll let you know how it goes!

Best wishes
Tania

On 24 August 2016 at 17:55, Anthony, Neil <nantho2 at emory.edu<mailto:nantho2 at emory.edu>> wrote:
Hi Tania,

Can I please ask what the z voxel size is?  I see in the test.xml that you have a voxel size of 0.235 in xy and z for id 0,1,2 in the ViewSetup tag.  When I create a xml from scratch here, setting xy to 0.235 and z to 1.5 (the smallest on the standard Picard 4D stage) I see each id in the xml has a size of “0.235 0.235 1.5”.

Stephan, I assume these are the voxel sizes used for the BDV.  Would I be correct in assuming some more and saying that if only one value is seen it’s used for xy and z?

…  interlude …  etc. etc.  lalalalala …  coffee etc.

I now see what the problem is.  Well, what some of the problems might be.

I did a manual transform around the y axis and manually overlapped the big bead.  See attached.  Is this how is should look?  If I have it laid out correctly I think you have a combination of not enough beads, combined with not enough overlap of physical volumes given the low number of beads.  On top of that I feel the small bead signal is so low that it’s approaching the camera noise*.

It is difficult as the big beads are so bright, and you might want to find a better sample to test with.  I’ve done some testing on spheroids that are really bright, and I had to swap my 100nm tetra spec beads for some really bright 1um green beads to be able to get good signal on the beads and not saturate the sample of interest.

I started with just little beads to figure out the basics, and I take just sub-res bead data to get my PSF output from the decon too.

Neil


* inspect the data stacks with orthogonal view and you’ll see the vertical lines running down the stack.  For this kind of noise you can run a few line scans across your beads and noise to get an idea of the intensity difference and then use remove outliers



From: OpenSPIM [mailto:openspim-bounces at openspim.org<mailto:openspim-bounces at openspim.org>] On Behalf Of Tania Mendonca
Sent: Friday, August 19, 2016 3:31 PM
To: Stephan.Preibisch at mdc-berlin.de<mailto:Stephan.Preibisch at mdc-berlin.de>
Cc: openspim at openspim.org<mailto:openspim at openspim.org>
Subject: Re: [OpenSPIM] Multiview registration - no inliers

Hi

Just tried it but it still doesn't work.

Thanks
Tania

On 19 August 2016 at 20:13, Stephan.Preibisch at mdc-berlin.de<mailto:Stephan.Preibisch at mdc-berlin.de> <Stephan.Preibisch at mdc-berlin.de<mailto:Stephan.Preibisch at mdc-berlin.de>> wrote:
Hi, I just think that in the beginning when you define your dataset the pixel sizes were wrong which potentially corrupts the downstream workflow. Try to start from the very beginning, and define a new dataset and manually put in the correct calibration. This might just work ...

All the best,
Stephan

Sent from my iPhone

On Aug 19, 2016, at 21:00, Tania Mendonca <tvmendonca1 at sheffield.ac.uk<mailto:tvmendonca1 at sheffield.ac.uk>> wrote:
Hi

I've just checked to make sure but yes I have corrected the voxel size on the image stacks and there shouldn't be a mistake in that. However, your reply gave me an idea and I did a quick test. I duplicated one of the images and ran the multiview reconstruction plugin for the two duplicates. It worked, so it makes me think that there's nothing wrong with the files themselves and you might be right about the issue being with my voxel depth. I've referred to SVI's Nyquist calculator to workout my step size, is there a recommended step size that people use?

Best wishes
Tania

On 19 August 2016 at 16:58, Stephan.Preibisch at mdc-berlin.de<mailto:Stephan.Preibisch at mdc-berlin.de> <Stephan.Preibisch at mdc-berlin.de<mailto:Stephan.Preibisch at mdc-berlin.de>> wrote:
Hi, my first guess would be that your calibration (pixel size is x,y and z) is wrong ... Can you make sure this is correct when you define your dataset?

All the best,
Stephan

On Aug 19, 2016, at 17:17, Tania Mendonca <tvmendonca1 at sheffield.ac.uk<mailto:tvmendonca1 at sheffield.ac.uk>> wrote:
Hi all

Hope your microscopes are aligned and treating you well! I'm sorry if I've missed this conversation before but I was wondering if any of you have had similar problems as me with trying to get the Fiji multiview reconstruction plugin to like my data.

I've tried a range of microspheres - different colours, different sizes and different concentrations but it never seems to find enough correspondences between views. I've been primarily experimenting with interactively detec​​ting interest points. I've tried the presets as well but with no luck. I've tried to systematically work through various combinations of sigma and threshold but always end up with 0 inliers. Registrations always terminates with the following error: "There are no connected tiles, cannot do an optimization. Quitting."

My home built system doesn't run on µManager and my images are acquired as .cxd files which I convert to .tifs before processing. I'm attaching some of my test data -https://drive.google.com/file/d/0B-he-_wcys_-aUtRZ0pidFhyUlk/view?usp=sharing
the sample here is 1µm yellow-green beads (1:4000 dilution) and 10µm beads (1:50 dilution) with the idea of using the 1µm beads to reconstruct the 10µm beads. I use the same objectives as on the OpenSPIM system (10X for illumination and 20X for detection), magnification on the system is 27.7X. The illumination here is 473nm and detection at 520nm.​​

A log file has been included in the zip file to give you an idea of what I've been trying.

I hope I'm just missing a trick somewhere. Any help would be greatly appreciated!

Best wishes
Tania
--
-------------------------
Tania Mendonca
Doctoral Research Student
University of Sheffield
S10 2TN
-------------------------
_______________________________________________
OpenSPIM mailing list
OpenSPIM at openspim.org<mailto:OpenSPIM at openspim.org>
http://openspim.org/mailman/listinfo/openspim



--
-------------------------
Tania Mendonca
Doctoral Research Student
University of Sheffield
S10 2TN
-------------------------



--
-------------------------
Tania Mendonca
Doctoral Research Student
University of Sheffield
S10 2TN
-------------------------

________________________________

This e-mail message (including any attachments) is for the sole use of
the intended recipient(s) and may contain confidential and privileged
information. If the reader of this message is not the intended
recipient, you are hereby notified that any dissemination, distribution
or copying of this message (including any attachments) is strictly
prohibited.

If you have received this message in error, please contact
the sender by reply e-mail message and destroy all copies of the
original message (including attachments).



--
-------------------------
Tania Mendonca
Doctoral Research Student
University of Sheffield